Any one has experince with Thermo's Trace GC Ultra-TSQ MS/MS? We had one installed recently, but is experiencing some high N2 signal (comparing to H2O signal, H2O/N2 ~2:1.2 to 1.3) during the turing process. The vacaum is very good. Anyone sense there my be a leaking? Thanks in advance.
-
- Posts: 586
- Joined: Mon Oct 04, 2004 3:00 am
What is the ratio of nitrogen to oxygen? If it about 4 to 1 you have an air leak. I suspect the ratio will be much larger, around 50 or 100 to1.
Do you have one of the Valco traps on the carrier in line? These often release a large quantity of nitrogen when first installed, and this lasts for a long time if the trap in not mounted vertically. If you have one of these traps try running at a 250:1 split for 20 minutes and see if the nitrogen goes down.
Do you have one of the Valco traps on the carrier in line? These often release a large quantity of nitrogen when first installed, and this lasts for a long time if the trap in not mounted vertically. If you have one of these traps try running at a 250:1 split for 20 minutes and see if the nitrogen goes down.
-
- Posts: 218
- Joined: Thu Mar 31, 2005 5:33 pm
Thank you very much, Ron.
The N:O ratio is around 3:1 consistently. There is a N-pretreated filter in line, but have it purged for quite a while.
The mystery is when it's operated at 1ml/min, N signal is at normal range, but when flow rate is increased to 2, N become abnormal. We have a thermo technican here struggling. The answer is likely to be "no problem".
The N:O ratio is around 3:1 consistently. There is a N-pretreated filter in line, but have it purged for quite a while.
The mystery is when it's operated at 1ml/min, N signal is at normal range, but when flow rate is increased to 2, N become abnormal. We have a thermo technican here struggling. The answer is likely to be "no problem".
3 posts
Page 1 of 1
Who is online
In total there are 32 users online :: 0 registered, 0 hidden and 32 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm
Users browsing this forum: No registered users and 32 guests
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm
Users browsing this forum: No registered users and 32 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science