Contaminated column
I had a ISTD contamination problem on my GC/MS, I have replaced the liner, septa, cut a small piece off of the injection side of the column, baked out the column, and done solvent rinses, but nothing seems to work. Does anyone have anysuggestions as to what I should try next. The ISTD is 1-Chloro,octane.
Thanks.
Hello!
If you have a autosample change sirnga and change vials.
And then put injector and column high temperature for a night (check the t max).
And we hope good!
Some questions first:
does the contamination appear as a peak, or as background ?
If you run only the GC-MS programme, with no injection, do you still see the contamination ?
If you do a dummy injection (i.e. nothing in the syringe) do you still see the contamination ?
If you inject pure solvent do you still see the contamination, and is it still there with different solvents ?
Yes the comtamination is a peak.
I am working on the no injection now, if I do an "air" injection yes it is still there.
Yes at solvent injection still shows the peak.
I am having a service tech come out tomorrow to address the issue, so I will let you guys know the out come of the problem for furture reference.
It was a durty liner?
Glass wool contaminated?
Sorry I never got back to this I have been slammed at work. Anyway the injection port ended up having to be replaced along with column. All the parts of my autosampler had to be cleaned as well.
Users browsing this forum: Semrush [Bot] and 13 guests
