Is anyone have the experience of farnesyl pyrophosphate?

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
As the title, I tried to analyze farnesyl pyrophosphate, mevalonate acid, geranyl diphosphate using LC/MS/MS. I met a lot of problems in the analysis.

1. the standards can hardly eluent from column under normal condition only after NH4HCO3 (pH 8.5) modifier added. Do anybody have idea what's the use for NH4HCO3?
2. when i tried to quantitate the farnesyl pyrophosphate, I found the signal decreasing when the concentration is lower around 2nM, while it works fine when the concentration is around 20nM. at first, I thought it maybe result from absorption on glass vials. After I compared with glass, plastic, de-activated glass, I can not say that. Do anyone have any idea about that?


Thank you very much!

One thing I would suggest is eliminating all metal possible. The pyrophosphates are horrible about sticking to stainless. I used neutral ammonium acetate rather than basic buffer and they eluted fine.

Best,
Marc
2 posts Page 1 of 1

Who is online

In total there are 26 users online :: 1 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Semrush [Bot] and 25 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry