Quantitative Analysis Of Organic Waste by Cappillary GC
Hy Im a student,and I need help in understanding the below procedure.
This is the experimental setup:
(1) Qualitative analysis of organic waste samples
Internal Standard is Ethyl Benzene
Test solvents include: methanol,Acetonitrile,Toluene and Acetone
1) Prepare each of the test solvents and waste samples by the addition of IS.This is achieved by the addition of 1 mls of IS to 1 ml of solvent.
2) Dilute each of the samples and test solvents with dichloromethane in a ratio of 50:50 (v/v)
3) Inject each test solvents and record their retention time's.Also inject the waste samples
4) Qualitatively determine which solvents are present in the waste samples.Use the chromatograms to determine the % composition.
My question here is how to determine the % composition?
(2) Quantitative analysis
Here they asked me to find the Response Factor (F) to find the concentration,of the unknown mixture.
concentration of std/conentration of IS=Fx area of std //area of IS.
My question is
how to find the concentration of std and concentration of IS if there concentrations are not given?
Should i calculate it,if so how?
Sanju Simon
Use area normalization
(individual area / total area) x 100 = % composition
For ISTD calculate RRF (relative respond factor)
(Ci / Ai) x (As / Cs)
Where: Ci = Concentration of individual component
Ai = Area of individual component
As = Area of ISTD
Cs = Concentration of ISTD
Hopefully this information can help you.
M Azmi
Well the concentration is actually given or rather can be calculated in 2 ways.
You know that you add 1ml. So if doing it accurately you can always look up the density for each solvent and calculate the final concentration. Should do for a student project.
Alternatively but a bit tricky if due to evapouration: add the IS first to a vessel (less volatile), put on a scale and zero it, then add 1ml of the solvent of intrest and you get the weight.
How do i do that?
For example take Methanol.It has a density of .79130 g/ml.I used 1 ml.Therefore the Mass i used will be .79130 x 1 =.79310g.
After finding the mass I used,whats the nxt step?
How do I express this in g/ml or in any other concentration terms?
Sanju Simon
Sanju,
If I understand your dilution scheme correctly, you have 0.79g/4ml total volume of sample or ~0.2g/ml. That gives you a response for that concentration of that solvent.
By the way, it should come out pretty close to the area % of that component relative to the other components if you are using an FID. Not exactly because you are not dealing with straight hydrocarbons, but close.
Best regards.
