LC/MS/MS...which of them is better?
4 posts Page 1 of 1
1- agilent 6410, multi mode esi/apci
2- applied biosystem api 2000, APCI
3- Finnigan TSQ-quantom access, APCI
4- waters tqd, z-source with apci
all of them are triple quadrupool since we are looking for quantitative analysis for narcotics traces in blood and urine
ur kind help is appreciated
thanx
Agilent 6410, i don't know the source robustness for your application...
API 2000, prefer to think about API 3200>OK
Quantum access> OK
TQD > OK
I'm interested of your feed back about Agilent if you test it to compleet my blog www.lc-ms-ms.com
Regards
The API 2000 is a bit "long in the tooth" these days and is their oldest model (may have support issues in the future).
Considering they've released the 3000, 3200, 4000, 5000 and 5500.
Think about going for a reconditioned higher spec API instrument.
Good judgment comes from bad experience, and a lot of that comes from bad judgment.
I fully accept JGK statement. You better look at ABI's 3200 or 4000 depending on your LLOQ & LLOD.