METROHM vario 940 - Sulfite and sulfate det.
Dear all,
Have you got any experience with determination SULFATE and SULFITE in sample from technologycal waters. I can use for this app. METROHM vario 940 with SeS suppresion. Can I use Metrosep A Supp 10 250/4 for this app? Have you got experience with standard stability ? eg Will not it decompose SO3 to SO4 during calibration ?
Thank you
BR
Sam
In order to separate and to make sure there is no decomposition you need to add about 1-2.5% acetone to your eluent. Also ASUPP 7 250 is the recommended column for this analysis.
-Doug Castrillo
dcastrillo@metrohmusa.com
Metrosep A Supp 10 also works.
The best stabilization of sulfite is adding 10 mmol/L formaldehyde to the sample /Diluent. Also 2% isopropanol showed to work.
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
Thank you,
I found IC-standard from Fluka (Sigma - Aldrich), 1000 μg/mL in 1% NaOH, do you have any experience with standard ?
Thank you
This standard is stable as it is. For dilution I would recommend to use one of the above mentioned stabilization methods. These dilutions should be renewed quite frequently.
Dr. Markus Laeubli
(retired)
Metrohm AG
9101 Herisau
Switzerland