Column Storage Question
I was running some hLH/hFSH fractions on my column (Superdex 75 - 10/300 GL) with an isocratic buffer (CH3CN/NH4HCO3) and when I was finished I decided to purge the system and column with milli-q water thinking that the worst that would happen would be I would have to re-equilibrate next time I used it. Manufacturer's website is kind of sketchy on the storage conditions (other than pH and chemical stability). Was this acceptable? Any forseeable problems with long-term storage with just de-ionized micro-filtered water? Link to manufacturer's specs below.
http://www1.gelifesciences.com/aptrix/u ... eid=165422
I believe I'd make up CH3CN - H2O flush solvent in the same ratio as in your buffer, and wash out all the carbonate, and leave it packed with that.
Thank you kindly sir
I would never leave my column sitting in water (even if purified/milli-Q).
If not properly flushed, it can lead to buffer precipitation. Flush with ONLY milliq water to get rid of the salt, completely. Follow this with the ACN/water (without buffer) flush and store in the same.
Long term storage in water can create bacteria, de-stabilize your stationary phase (resulting in longer re-equilibration times, and above all can hydrolyze your stationary phase.
Always aim to store in a 100% ACN if long term. But do not forget to flush with ACN/water prior to running your salted mobile phase.
In total there are 18 users online :: 2 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Users browsing this forum: Bing [Bot], Semrush [Bot] and 16 guests