ELSD Signal Unstable
Hey, I have been suffering a ELSD signal unstable problems for a while. Here are the conditions we are using
Purity by HPLC
HPLC System: Agilent 1100/1200 HPLC System
Flow Rate: 0.5 mL/min
Mobile Phase A: 0.04% TFA in Acetonitrile
Mobile Phase B: 0.04% TFA in Water
Column Temperature: 50°C ± 5°C
Autosampler Temperature: Ambient
Run Time: 16 minutes
Detection System: Shimazu ELSD-LT II
Temperature: 75°C
Gas Pressure: 50 psi
Detector Environment: 50HƵ
However, the %RSD for system suitability is out of 10%. Could anyone have an idea about solving the problem. Appreciate it!
What are you analyzing? Make sure that your compound is not volatile, otherwise at lower concentrations you will lose linearity. Try to reduce temperature to 50 or even 35 degrees Celsius.
Vlad Orlovsky
HELIX Chromatography
www.helixchrom.com
To follow up with what was said previously. I think the ELSD temperature is too high. Try nearer 30C to see if that works better.
Kevin
