MS2 Spectra containing fragment mass same as precursor ion
Hi All,
I've some MSMS data (MGF files) for which I've found that there is a fragment mass almost equal to the mass of the precursor ion with highest abundance among the msms peaks, and some masses greater than the precursor. Is this something uncommon suggesting that the precursor is not adequately fragmented and formed adduct masses larger than the precursor?
Secondly, for the mass 457.5577859, can it result due to a loss of water molecule?
I'm enclosing part of the ms2 spectra with fragments sorted by their abundances. The third colums shows the relative abundance compared to the most abundant ms2 peak.
Code: Select all
BEGIN IONS
TITLE=C31 "controllerType=0 controllerNumber=1 scan=81"
RTINSECONDS=81.0765
PEPMASS=463.220001220703 2351077.0
CHARGE=3+
463.8945914   750950.5   100
457.5577859   266115.2813   35.43712685
360.6910986   194093.5781   25.84638776
457.2216554   173094.5625   23.05006289
646.3289232   146484.1875   19.50650376
668.2983467   109999.5469   14.6480423
439.5647401   107747.8906   14.34820146
577.7872263   98013.15625   13.05187975
577.2840984   95289.23438   12.68914987
645.8266372   92351   12.29788115
361.1934364   87724.55469   11.68180255
439.2308563   62405.16797   8.310157323
463.5589927   60907.16406   8.110676278
464.2267778   54563.89453   7.265977522
451.8826723   47210.08203   6.286710247
451.556398   45388.90234   6.044193638
669.2975931   41486.14063   5.524484054
646.8317955   39324.73828   5.236661841
425.219467   37794.64063   5.032907046
caveat: peptides aren't my thing
(1) the precursor mass appearing as the major fragment isn't uncommon, and can indicate that there wasn't enough collision energy
(2) in your case, the major fragment is actually an isotope peak of the precursor mass, not the actual precursor?? If it's not the most abundant isotope peak, then it's worth taking a look at how it happened, which might depend on your instrument. Some instruments (e.g. ion traps) can apply fragmentation energy to a particular bandwidth of masses, which may not be the same bandwidth as was selected for MS2, so it's possible to collect, say, 3 isotope peaks, but fragment only two of them efficiently. Also, if you're really unlucky, in any instrument, you can find a contaminant with nearly the same mass as the expected product, but which is much tougher and won't fragment. In this scenario your expected product fragments, leaving a near-isobaric peak from the unfragmented contaminant. I don't think this has happened in your case.
(3) It's quite normal to get fragments "bigger" than the precursor when working with multiply charged ions, because you can lose charge as well as mass. The mass:charge ratio of a singly-charged fragment from your triply-charged precursor could reach nearly three times the value of the precursor (i.e. your original precursor has a mass of 463 * 3 = 1389, so if you lose two charges and a bit of mass, you could theoretically get a singly charged fragment showing up at over 1300!).
(4) yes, a loss of 6 from a triply-charged ion could well indicate a loss of 18, water. But because you're looking at a peptide, a loss of 17 (ammonia) is also likely. You need to check which isotope peak you're looking at, to make sure the loss is really 6 and not 5.67; it's very easy to end up looking at the mass difference between different isotope peaks of a precursor ion and a fragment when there are multiple peaks with good intensity.
Thanks a ton
Users browsing this forum: Bing [Bot] and 5 guests